I generate a sine wave with the command line utility (for example: "generate 1 0.01 100 sine").
Now I measure the outgoing signal with a oscilloscope and see that the sine wave is not in the middle of zero.
I change the calibration coefficient "BE_CH1_DC_offs" with the calib -w command but the offset of the signal is not changed.
What do I wrong?
Analog back-end calibration
- redpitaya
- Site Admin
- Posts: 912
- Joined: Wed Mar 26, 2014 7:04 pm
Re: Analog back-end calibration
Hi,
You want to calibrate free applications?
Best, RP
You want to calibrate free applications?
Best, RP
-
- Posts: 15
- Joined: Mon Feb 22, 2016 11:47 am
Re: Analog back-end calibration
Hi,
I don't use a application. I generate the signal with the command line.
Is the calibration only available if I generate a signal with a app?
I don't use a application. I generate the signal with the command line.
Is the calibration only available if I generate a signal with a app?
-
- Posts: 31
- Joined: Mon Mar 28, 2016 1:10 pm
Re: Analog back-end calibration
hi,
if you have a reference to calibrate against,
and use desktop tools like matlab or python,
it will be easiest for you to calibrate in post-processing (on desktop).
Otherwise you will find yourself reading source code of the FPGA.
if you have a reference to calibrate against,
and use desktop tools like matlab or python,
it will be easiest for you to calibrate in post-processing (on desktop).
Otherwise you will find yourself reading source code of the FPGA.
-
- Posts: 1441
- Joined: Sat Jun 07, 2014 12:49 pm
- Location: Königswinter
Re: Analog back-end calibration
Compensating signals for calibration is all done outside the FPGA at the moment.Otherwise you will find yourself reading source code of the FPGA.
Your problem is that the generate tool is not programmed to use the calibration values. I never tire of recommending edgo's calibration-aware tools.I don't use a application. I generate the signal with the command line.
Is the calibration only available if I generate a signal with a app?
-
- Posts: 15
- Joined: Mon Feb 22, 2016 11:47 am
Re: Analog back-end calibration
I found a solution to compensate the offset of the analog back end. I manipulate the fpga registers 0x40200004 for channel 1 and 0x40200024 for channel 2 with the monitor tool.
My next problem is that I need to to this every time I power up red pitaya.
Is it possible to to start a shell script after red pitaya finished the boot process.
I insert my command in the rcS file in opt/etc/init.d location but this dosen't work.
(/etc/init.d/rcS: /opt/etc/init.d/rcS: line 22: monitor: not found)
My next problem is that I need to to this every time I power up red pitaya.
Is it possible to to start a shell script after red pitaya finished the boot process.
I insert my command in the rcS file in opt/etc/init.d location but this dosen't work.
(/etc/init.d/rcS: /opt/etc/init.d/rcS: line 22: monitor: not found)
-
- Posts: 15
- Joined: Mon Feb 22, 2016 11:47 am
Re: Analog back-end calibration
It was a failure of myself.
I forgot to give the path to execute command.
opt/bin/monitor 0x...
I forgot to give the path to execute command.
opt/bin/monitor 0x...