Hey there, I'm having trouble setting the trigger level higher than 1V
"ACQ:TRIG:LEV 1000 mV" works perfectly fine but "ACQ:TRIG:LEV 1001 mV" returns an error.
Anybody know what's causing this?
Can't set trigger level above 1V.
-
- Posts: 1441
- Joined: Sat Jun 07, 2014 12:49 pm
- Location: Königswinter
Re: Can't set trigger level above 1V.
The nominal measuring range of the Red Pitaya in LV input configuration is +-1V. The samples are capped to that range, so what would be the sense in allowing a higher trigger level?
-
- Posts: 2
- Joined: Wed Mar 16, 2016 12:02 pm
Re: Can't set trigger level above 1V.
What about in HV mode? how high should I be able to measure then?
I fixed my problem. I had SOUR1 in HV mode and was wondering why I couldn't set the trigger level higher but SOUR2 was still in LV. When both are in HV mode, I can set it higher.
For anybody googling this in the future:
I fixed my problem. I had SOUR1 in HV mode and was wondering why I couldn't set the trigger level higher but SOUR2 was still in LV. When both are in HV mode, I can set it higher.
For anybody googling this in the future:
Code: Select all
ACQ:SOUR1:GAIN HV
ACQ:SOUR2:GAIN HV
ACQ:TRIG:LEV 5000 mV
Last edited by colmmoore on Tue Jun 28, 2016 3:00 pm, edited 1 time in total.
-
- Posts: 1441
- Joined: Sat Jun 07, 2014 12:49 pm
- Location: Königswinter
Re: Can't set trigger level above 1V.
+-20V is the HV input range.
edit:
Now that I examine setting of the trigger level, it is a bit odd how it's implemented to set both channels simultaneously. If you feel that this could be handled better, I'd totally agree.
edit:
Now that I examine setting of the trigger level, it is a bit odd how it's implemented to set both channels simultaneously. If you feel that this could be handled better, I'd totally agree.