Manual calibration

Posted: Sun Oct 30, 2022 3:43 pm
by fgeorgy
Hi,

I'm trying to calibrate my RP but since I don't have a 0.5V reference available and I never use the LV range, I was wondering if it is possible to only calibrate the HV range.

I tried with the manual calibration but I can't get to something usable and am not sure of doing it right.

Does anybody know how to use the manual calibration procedure?

Re: Manual calibration

Posted: Wed Nov 23, 2022 10:34 pm
by redpitaya
Hey,

Sorry for the late reply.

You can use some sort of a resistor divider to achive the wanted voltage (0.5 Volts) if you have a higher voltage source. That could solve your problem.

The calibration currently does not support calibration for only one voltage range.