Hey pitaya users,
i am about to calibrate my pitaya and i don't fully understand the calibration vector you get with the "calib -r -v" command.
I understand the offset calibration as it is described in the user manual but what about the "full scale coefficients"?
FE_CH1_FS_G_HI = 45870551 # Front-End Channel1 full scale coefficient for High gain (LV) jumper configuration.
FE_CH2_FS_G_HI = 45870551 # Front-End Channel2 full scale coefficient for High gain (LV) jumper configuration.
What are the units of these numbers? I expect something like "samples per volts" but the number is much larger.
So if you acquire a measured number of samples with the "acquire" command, how do you compute a measured voltage out of it with this coefficient and the offset?
I would really appreciate any help on this.
Regards
units of calibration coefficients?
-
- Posts: 1441
- Joined: Sat Jun 07, 2014 12:49 pm
- Location: Königswinter
Re: units of calibration coefficients?
These are the sections of code that transform samples to voltages by way of the calibration values:
(pieced together from the scope app's worker.c and fpga.c , see there for the missing constants, types and parameters)
The unit of the calibration value 'full scale coefficient' is V/cnt , but there are some constants factored into it. This is why its magnitude seems kind of arbitrary.
Code: Select all
float sample_to_voltage(int sample) {
float voltage, ch1_max_adc_v;
uint32_t fe_fsg1;
fe_fsg1 = (curr_params[GAIN_CH1].value == 0) ? rp_calib_params->fe_ch1_fs_g_hi
: rp_calib_params->fe_ch1_fs_g_lo;
ch1_max_adc_v = osc_fpga_calc_adc_max_v(fe_fsg1, (int)curr_params[PRB_ATT_CH1].value);
voltage = osc_fpga_cnv_cnt_to_v(sample, ch1_max_adc_v, rp_calib_params->fe_ch1_dc_offs, 0);
return voltage;
}
/*----------------------------------------------------------------------------*/
/**
* @brief Converts ADC counts to voltage [V]
*
* Function is used to publish captured signal data to external world in user units.
* Calculation is based on maximal voltage, which can be applied on ADC inputs and
* calibrated and user defined DC offsets.
*
* @param[in] cnts Captured Signal Value, expressed in ADC counts
* @param[in] adc_max_v Maximal ADC voltage, specified in [V]
* @param[in] calib_dc_off Calibrated DC offset, specified in ADC counts
* @param[in] user_dc_off User specified DC offset, specified in [V]
* @retval float Signal Value, expressed in user units [V]
*/
float osc_fpga_cnv_cnt_to_v(int cnts, float adc_max_v,
int calib_dc_off, float user_dc_off)
{
int m;
float ret_val;
/* check sign */
if(cnts & (1<<(c_osc_fpga_adc_bits-1))) {
/* negative number */
m = -1 *((cnts ^ ((1<<c_osc_fpga_adc_bits)-1)) + 1);
} else {
/* positive number */
m = cnts;
}
/* adopt ADC count with calibrated DC offset */
m += calib_dc_off;
/* map ADC counts into user units */
if(m < (-1 * (1<<(c_osc_fpga_adc_bits-1))))
m = (-1 * (1<<(c_osc_fpga_adc_bits-1)));
else if(m > (1<<(c_osc_fpga_adc_bits-1)))
m = (1<<(c_osc_fpga_adc_bits-1));
ret_val = (m * adc_max_v /
(float)(1<<(c_osc_fpga_adc_bits-1)));
/* and adopt the calculation with user specified DC offset */
ret_val += user_dc_off;
return ret_val;
}
/*----------------------------------------------------------------------------*/
/**
* @brief Calculates maximum [V] in respect to set & calibration parameters
*
* Function is used to calculate the maximal voltage which can be applied on ADC inputs.
* This calculation is based on calibrated Front End Full Scale Gain setting and configured
* Probe Attenuation.
*
* @param[in] fe_gain_fs Front End Full Scale Gain
* @param[in] probe_att Probe attenuation
* @retval float Maximum voltage, expressed in [V]
*/
float osc_fpga_calc_adc_max_v(uint32_t fe_gain_fs, int probe_att)
{
float max_adc_v;
int probe_att_fact = (probe_att > 0) ? 10 : 1;
max_adc_v =
fe_gain_fs/(float)((uint64_t)1<<32) * 100 * (probe_att_fact);
return max_adc_v;
}
The unit of the calibration value 'full scale coefficient' is V/cnt , but there are some constants factored into it. This is why its magnitude seems kind of arbitrary.
-