I have a unit configured with IN1 and IN2 set to HV and probes set to 10x.
Using the Oscilloscope I have connected IN1 to 5V source 1s ON/ 1 Sec OFF and the crocodile connector of this probe to the - of the source. The scope displayes -5.14.
Connecting the same source to IN2 produces a similar signal..
I executed a factory reset on the calibration. But the problem persists.
See attached screen.
I tried auto calibration, but the step where it wants me to connect to a 5V source doesn't work because it doesn't seem detect the 5V. Next to the input field with the reference value, the indicator remains red.
Do I have a hardware problem?
Stemlab 125-14 hardware problem??
-
- Posts: 4
- Joined: Fri Jun 03, 2016 9:15 am
Re: Stemlab 125-14 hardware problem??
See screen
-
- Posts: 4
- Joined: Fri Jun 03, 2016 9:15 am
Re: Stemlab 125-14 hardware problem??
Attachements seem to fail on this forum....
Bummer, so no replies nor support..
Bummer, so no replies nor support..
-
- Posts: 50
- Joined: Sun Mar 11, 2018 6:03 pm
Re: Stemlab 125-14 hardware problem??
You don't have a hardware problem - you should do the calibration properly. No such thing as auto-calibration.