Page 1 of 1
Sensetivity of the ADC input
Posted: Tue Feb 08, 2022 7:15 am
by ASDFG123
Hi, what is the level of sensetivity of the 125-14 and 122-16 boards? Conditions is the sine signal 10-15 MHz, both boards terminated to 50 Ohms (125-14 externally), both have simple osciloscope flash program. I did not noticed any advantage of the 122-16 board, actually 125-14 is better. Both see signal from 0.5 - 1 mV RMS, but 14 bit board has wider dynamic range.
Is this result that should ber or I have some mistake?
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 8:44 am
by Yevgeniy
Hi!
I have the same question about sensivety of SDRlab 122-16.
We did tests and the results are below.
Input signal: 200mV RMS
https://drive.google.com/file/d/1DLp82E ... sp=sharing
Input signal: 370µV RMS
https://drive.google.com/file/d/1xk4VnW ... sp=sharing
Input signal: 120µV RMS
https://drive.google.com/file/d/19EGX9O ... sp=sharing
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 8:55 am
by pavel
ASDFG123 wrote: ↑Tue Feb 08, 2022 7:15 am
Both see signal from 0.5 - 1 mV RMS, but 14 bit board has wider dynamic range.
On the 122-16 board, changing the position of the jumper connected to the SENSE pin of the ADC chip increases the dynamic range by 6 dB without decreasing the sensitivity.
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 9:21 am
by Yevgeniy
Hi Pavel,
I couldn't found any jumpers on SDRlab-122-16.
Maybe you mean STEMlab-125-14?
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 9:42 am
by pavel
Yevgeniy wrote: ↑Thu Feb 10, 2022 9:21 am
I couldn't found any jumpers on SDRlab-122-16.
These are solder jumpers shown as 0 Ohm resistors R15 and R16 on page 3 of the 122-16 schematics.
To increase the dynamic range, R15 should be unsoldered and R16 should be soldered.
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 10:09 am
by Yevgeniy
The SENSE pin change input range from 0,5V to 1V. It's clear.
But it's not solve problem because we does couldn't to digitize signal below 200 µV.
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 10:37 am
by pavel
Yes, my comment was about the dynamic range. On the 122-16 board, it can be increased.
If your signal is periodic and your application does not require 60 MHz bandwidth, then I think the sensitivity can be increased by downsampling.
Re: Sensetivity of the ADC input
Posted: Thu Feb 10, 2022 11:05 am
by pavel
I have just done a few quick checks and calculations.
The LTC2185 ADC on the 122-16 board has an SNR of around 76.8dB for the +/-1V input range.
The ADC noise amplitude can be calculated as
1V / (10^(76.8dB / 20)) = 145uV
This seems to match what you observe quite well.
edit: just noticed that the WBC4-1TL transformer has a turns ratio of 1:2 and it should add a factor of 2 to the noise amplitude calculation at the board input (72uV).
Re: Sensetivity of the ADC input
Posted: Mon Feb 28, 2022 10:23 am
by redpitaya