Can I safely overload LV mode [Stemlab 125-14 DK]

Placement, modules, components and accessories; the ones that exist and the the nice-to-be's
Post Reply
MaxHe
Posts: 4
Joined: Tue Jun 27, 2023 11:01 am

Can I safely overload LV mode [Stemlab 125-14 DK]

Post by MaxHe » Tue Jun 27, 2023 12:43 pm

Hello,

Is it possible to measure a signal which is potentially between ±5V (Sensor output with integrated amplifier) with the ±1V input range. I want a higher resolution, and anything above 1V is not really of interest for my application. Just want to check that I do not damage the board.
:D
Best regards

Maximilian

User avatar
redpitaya
Site Admin
Posts: 912
Joined: Wed Mar 26, 2014 7:04 pm

Re: Can I safely overload LV mode [Stemlab 125-14 DK]

Post by redpitaya » Wed Jul 05, 2023 1:29 pm

Hello Maximilian,

Thank you for contacting us through the forum.

The fast analog inputs can handle a maximum of +-30 V (this is the maximum the voltage protection can handle).

The board should not be damaged by the 5 V signal on LV jumper settings. However, we do not recommend using it outside the specified voltage ranges (LV +-1V, HV +-20V).